Oh no, can't use my Bergeon micrometre! Help needed!


VWatchie

Recommended Posts

image.png.e25961afb8df7d27c9d228f960ca476f.png

image.thumb.png.c1e5de6a24e89c5d7e307e45dca3de78.png

image.thumb.png.8a1b9a2a8076a39210efac4e7bc7d40a.png

image.png.9da7261626780a97c34d438661f200fb.png

image.png.570aeb4bf77de5ba2864b12ad968c640.png

Some time ago I bought a Bergeon micrometre (and a Seitz jewelling tool) to be able to replace damaged jewels. The micrometre works perfectly down to the exact hundredth of a millimetre but trying to measure my first jewel I bumped into a very unexpected problem; the height adjustable round table fouls the spindle (circled in red).

I don’t think, or rather I hope, that the micrometre isn’t out of order and can be adjusted so that the spindle does not bump into the table. But searching high and low I just can’t find any documentation or user manual for this type of micrometre. Being a hobbyist having spent a small fortune on these tools, this situation is very frustrating.

So, the question is, what can I do to adjust my micrometre? As I have no clue what the various parts of the micrometre are named, I have given the various parts in the pictures a letter (in yellow) so that if anyone who knows how do this (and takes pity on me :unsure:) more easily can guide me.

Keeping my fingers crossed. Hard!

image.png

Edited by VWatchie
Tried to remove the last picture at the bottom of the post
Link to comment
Share on other sites

Hi Watchie   from what I can see  and not ever having used one of these, the table is for supporting parts being measured and in no other way interferes with the actual measurement. So by fitting a shim or disc with a lesser diameter to the table top to allow the arm to clear the table and support the piece being measured  may solve the problem . 

Link to comment
Share on other sites

it looks to me that if you loosen the two screws A & B then move the micrometer away from the table, move shaft C, which is now loose to the middle of the table E.

Tighten the two screws A & B 

You should the have the meeting point of the two shafts in the middle of the table and both point will be able to meet each other.

Edited by Alaskamick
modify
  • Like 4
Link to comment
Share on other sites

16 minutes ago, watchweasol said:

Hi Watchie   from what I can see  and not ever having used one of these, the table is for supporting parts being measured and in no other way interferes with the actual measurement. So by fitting a shim or disc with a lesser diameter to the table top to allow the arm to clear the table and support the piece being measured  may solve the problem . 

Thanks for the suggestion and the thought had occurred to me (lo and behold ;)) but I really don't think that's how it was designed and meant to be used. @Mark uses a similar micrometer, and here you can see how the table is supposed to support the jewel while being measured. I really don't want to add an extra disc/shim as it will introduce another piece to handle, and I'm not even sure how well it would work. Well, as a last resort I guess it could be OK.

Edited by VWatchie
Link to comment
Share on other sites

11 minutes ago, Alaskamick said:

it looks to me that if you loosen the two screws A & B then move the micrometer away from the table, move shaft C to the middle of the table E. you   should the have the meeting point of the two shafts in the middle of the table and both point will be able to meet each other.

Yes, I've considered that as an option too, but then the scale wouldn't be correct. I guess if I move it a millimeter I could always subtract a millimeter from the reading, but sooner or later I'm going to forget and mess it up. Anyway, thanks for the suggestion and I might try it if a better (or more correct) way doesn't come up.

Link to comment
Share on other sites

You shouldn't have any trouble. Once you have the two spindle flats touching in the centre of the table all you have to do is adjust the location of the micrometer scale, as with any micrometer they can be altered easily. It is normal practice.

 

I didn't notice that the ends of the two spindles were Flat.

Just bring them together after loosening the two screws earlier mentioned, then you adjust the scale by loosening collar H, line the micrometer up where you can see it easily then retighten H.

Screw L may lock the micrometer to stop it turning so that will need ti be loosened a little

 

Get the LH side in the right place first with the flat in line with the centre of the table then move to the micrometer side and start altering that.

Edited by Alaskamick
Link to comment
Share on other sites

if i am correct you should be able to adjust the left pin or what to call it . By unscrewing  the A and B screws . never mind the right pin isn't zero . That you can adjust later.  The big screw on the ratchet is only for  taking the click mechanism  apart inside the handle .  To adjust the zero on the scale loosen the big hex nut on the  back of the scale . And turn the scale to zero .  That way you will get clearance  for the table. 

  • Like 1
  • Thanks 1
Link to comment
Share on other sites

41 minutes ago, Alaskamick said:

it looks to me that if you loosen the two screws A & B then move the micrometer away from the table, move shaft C, which is now loose to the middle of the table E.

Tighten the two screws A & B 

You should the have the meeting point of the two shafts in the middle of the table and both point will be able to meet each other.

I believe after he moves C, he'll have to adjust H,L and K to re-establish his zero point, which is C and F touching face to face. EDIT: @rogart63 has it correct!

Edited by FLwatchguy73
corrected info
  • Like 1
Link to comment
Share on other sites

2 minutes ago, FLwatchguy73 said:

I believe after he moves C, he'll have to adjust H,L and K to re-establish his zero point, which is C and F touching face to face. EDIT: @rogart63 has it correct!

Maybe not 100% . The K can be used to align the zero to the scale to zero .  It no harm in trying . As long as you  arne't loosening tha screws to much.  Look at a  normal micrometer . They work the same way . 

  • Thanks 1
Link to comment
Share on other sites

You can definitely move the fixed anvil toward the mic head (to the right). Just enough that you clear the table. On one of mine there is only  maybe 0.10mm clearance, but that's fine. Once you've done that show us what it's doing on the scale and we can advise.

 

The collar H is to tighten a collet that takes up the play on the spindle. You won't need to touch that.

 

 

  • Thanks 1
Link to comment
Share on other sites

image.thumb.png.9490c5b72ff2d70760b53c35141321de.png

image.thumb.png.7b08bfcf78e130e5789865e90193be84.png

image.png.eda5ca316c2eaf82313ff1cec55c9cbe.png

Thanks guys! :thumbsu: All your support really makes me warm at heart! :wub:

As you can see, there has been some progress. (being someone who barely held a screwdriver in hands before the age of 55 I feel pretty proud of myself). Anyway, can't get the scale right. I loosened the big hex nut (even removed it as you can see) but the nut seems to be fixed to the screw, and no matter how I fiddle with it nothing happens. I guess something else has to be done to get it right...

Keeping my fingers crossed you can help me again!

Link to comment
Share on other sites

Put the screw/nut back on, you don't need to move the thimble (the barrel with the graduations 0-99). For future reference, it's held in place with that screw/nut, but is fitted on a taper and needs some coaxing to get off and move around.

 

If you loosen screw K you can move the inner sleeve, it will just slide around. Get it where you want it and tighten K and you're done.

  • Thanks 1
Link to comment
Share on other sites

The nut you have undone is only part of the micrometer mechanism, nothing to do with the mounting of the unit.

As I said in my earlier post try loosening screw L which I think locks it in place. also try undoing the  Ring H.

Something frees it up and allows it to move back and forth

Edited by Alaskamick
Link to comment
Share on other sites

image.thumb.png.a3d82b348f2a2569683e21a6c070c616.png
Mission accomplished. The Eagle has landed!

image.png.b9dd3e3da90e2ecf1b0cd9e7e62bec0b.png

So, I’m all good to go! Just can’t thank you enough guys! I know that for you seasoned repairers this is nothing special, but me, well, I feel like I’ve successfully taken off in a rocket! :lol:

For those of you finding this thread in the future here’s some additional info that I learned on the way.

After having removed screw A and B the spindle C just wouldn’t move pulling it by hand. So, I tapped on it ever so lightly horizontally in the direction towards spindle F and it moved in increments. To tap I used the plastic part of the hammer in the picture.

I also needed to rotate the spindles C and F slightly. I did this by grabbing on to the thick part of the left spindle C (after having tapped it out a bit) with pliers with a bit of transparent and though plastic film (zip bag) in between the pliers and the spindle. To rotate the right spindle F I simply used a small BAHCO adjustable wrench.

Loosening screw K as @nickelsilver and  @rogart63 point out is indeed the way to get the scale back to normal! The inner sleeve can then slide as well as rotate without affecting the thimble (the barrel with the graduations 0-99).

I’m really thrilled to own this beautiful precision tool! It gives me the same feeling that I get from looking at a really fine watch movement.

So again, a really big thank you to all of you! :thumbsu:
 

  • Like 5
Link to comment
Share on other sites

Join the conversation

You can post now and register later. If you have an account, sign in now to post with your account.
Note: Your post will require moderator approval before it will be visible.

Guest
Reply to this topic...

×   Pasted as rich text.   Restore formatting

  Only 75 emoji are allowed.

×   Your link has been automatically embedded.   Display as a link instead

×   Your previous content has been restored.   Clear editor

×   You cannot paste images directly. Upload or insert images from URL.



  • Similar Content

  • Recently Browsing

    • No registered users viewing this page.
  • Topics

  • Posts

    • The post below contains the link. If you don't already have a discord account it will take you to the registration screen.  Registration is free.   https://www.watchrepairtalk.com/topic/31653-mark/?do=findComment&comment=279066
    • HWGIKE#57 Valex FEF 190 15 jewels Swiss lever full service and repair This one was waiting for a balance staff replacement in my cabinet parts and case cleaned up with a new balance staff and a 4th wheel as the original 4th wheel had a broken pivot for the off center second hand. I never attempted a balance staff replacement before however I received a Bergeon Molfres (i was hunting it for about 2 years) and with the help of it I managed to remove the old staff and riveted the new one in. It also received a new MS, crystal and the hole for the MS arbor was also tightened. With the new MS now it has an acceptable performance meaning that the amplitude goes up to 280 fully wound, has an acceptable beat error and I have the two nice lines but only dial up, dial down is not as nice and I could not figure out as why. I have the two lines but the amplitude is dropping to around 230 and the lines are a bit hairy. Both dial up and dial down the lines just go up and down without seemingly any pattern. I cleaned the movement two times, and then a 3rd time pegged out the main plate and train bridge holes but made no change. Both the HS collet and the roller table was too lose on the new staff... I did not count how many times I took the balance cock off to sort out the HS collet, the roller table and the beat error, somebody before me also shortened the HS by pushing it out a bit and it seems every time somebody is messing with the end of the HS the protruding bit is most of the time twisted bent etc. This one was probably one of the most challenging repair and service. I might take the new MS out and clean it lubricate it as I just pushed the new one in to the barrel from the retaining ring. Plus started to re-read the theory of the escapement and how to analyse the graph on the timing machine: Greiner Chronografic Record manual. I am also thinking to put the watch on a 24 hour long run with the eTimer SW it once helped me to figure out what was wrong with a watch. There is an interesting part of the Greiner record manual talking about the pallets and the end shake of the balance and pallet staff. Maybe this is my issue? Who could that possibly identify? After a few years now I am still without a clue how could watchmakers make parts I can only see with my microscope or how could/can they carry out complicated services impossible to do.. real magic..... .... ..... before I sent this post while the pics were uploading I had an idea, i was browsing the possible outcomes on the timing machine I had one for magnetism..... so I demagnetized the movement and it is not hairy now.... two really nice lines 0.2 ms beat error still a bit wavy, but a lot lot better..... argh....  
    • Hi there, welcome here.  
    • yes the advertising revenue should generate money. The question is how much money? Then as far as the cost of the website goes that's relatively easy to determine? all you would have to do to grasp costs and profitability would be to go to the link below and you can actually get a website for free try it out for free I believe you get no advertising initially. They also talk about that they'll help you out they have marketing tools and some sort of paid subscription or something. So I guess were shopping for a whatever just ask them what would a maybe could use this one as an example in other words it's going to look basically identical to this is going to have advertising a paid subscriptions what's it going to cost? After all they want to sell or give us a message board like this they should bill answer the questions as they're the people who did the software for this. Yes they really said you can have a free discussion group at least to start. https://invisioncommunity.com/ I was curious about the monthly supporter thing where exactly do we find that on this message board? A quick search I'm not finding it so obviously I'm not looking in the right place?
    • Help me out here, but with all the advertising on this site (which I don't mind) wouldn't it pay for itself or even make money for the owner ???
×
×
  • Create New...