Posted

Hello,

I am fairly new to watchrepair, having successfully ”repaired” 5-6 watches. I am currently working on a faulty Girard Perregaux Gyromatic and ran into a broken balance for the first time. I ordered a new one (complete balance) but probably didn’t install it perfectly. 
Does any one know just by the chaos of the timegrapher what the problem could be?

thanks in advance

:)

IMG_2968.jpeg

IMG_2970.jpeg

Posted

I am also fairly new, but I'll take a stab based on the experience I have. This is the checklist I'd go over if it were on my bench:

  • Demagnetize
  • Check the impulse jewel for chips and whether it's loose
  • Check end shake on the balance jewels (low amplitude)
  • Check to see if the balance runs true (doesn't wobble when looking from the side) and isn't contacting anything
  • Check that the hairspring is seated properly, is in a parallel and flat plane to the balance wheel, and is not contacting anything

And then go from there. Keep us posted!

Posted

I did some minor adjustments which at least made the readings less chaotic. However, it still needs some work. Thanks for the tips, I will continue troubleshooting 🙂

IMG_2975.jpeg

I attached a picture of the movement. 
:)

IMG_2976.jpeg

I have now tried to demagnetize the watch, but with no real success.. I will contiue to look for problems 😕

if anyone has any other tips, please let me know. Is it worth it to tackle the beat error? Or should I focus finding the other problem/problems first?

IMG_2978.jpeg

Posted
1 hour ago, watchweasol said:

Hi the list proposed by manskirtbrew is a good comprehensive list to start with, follow that .

Pictures of the movement would be appreciated.

I agree with the above. Close up pics of the balance at rest, from the top and side would be useful. 

Posted
1 hour ago, Nucejoe said:

Regulator arm seems not fully snapped in place.

Well spotted @Nucejoe

You can see it not seated properly at the arrow.

As a result the hairspring does not sit between the regulator pins correctly. In the bottom pic you can see the hairspring going all the way through the pins, where it should be hidden by the boot for a portion

image.png.c38a3e2b97f755fffcf562cf40d06136.png

image.png.21f70ae07282d2917b52d81b139a7a76.png

 

 

 

  • Like 2
  • Thanks 1
Posted
1 hour ago, RichardHarris123 said:

What effect will that have? 

Wouldn't make regulating more precise, makes a mess of pin-boot slot.

 The oscilator seems out of beat. 

Farnkly this watch needs all adjustments listed by girls skirt brew unless your are Irish. lol

Rgds

 

 

  • Thanks 1
Posted

Thank you guys!

I thought it looked weird on my original inspection, but as I bought the complete balance and cock NOS I didn't give it to much thought.

I assume that I will have to disassemble the regulator completely?

I will try to do it later this afternoon, I'll keep you posted!

Thank you once, again.

Posted
1 hour ago, Nilbobaggins said:

... but as I bought the complete balance and cock NOS I didn't give it to much thought.

I assume that I will have to disassemble the regulator completely?...
 

 

No, just put the cock balance up on a flat surface, lift the balance alittle and hold it above the cock,then press the cock so the regulator will 'click' on it's place.

The whole thing is 'NOS', does it mean that You didn't dissassemble the incablok and clean/oil the bearings in the setting?

Another thing - it is not so right to replace the cock, especially if the original one is present. The correct thing is to replace only the balance assembly. If replacing the cock, must be checked if the balance stays upright and if the axial free play is correct.

  • Thanks 1
Posted
9 minutes ago, nevenbekriev said:

No, just put the cock balance up on a flat surface, lift the balance alittle and hold it above the cock,then press the cock so the regulator will 'click' on it's place.

The whole thing is 'NOS', does it mean that You didn't dissassemble the incablok and clean/oil the bearings in the setting?

Another thing - it is not so right to replace the cock, especially if the original one is present. The correct thing is to replace only the balance assembly. If replacing the cock, must be checked if the balance stays upright and if the axial free play is correct.

Okay, thank you for the advice!

The original cock was raided/destroyed so it was unusable sadly. However everything seems to line up "correctly", and Yes - I oiled and cleaned the cap-jewels. 🙂

  • Like 1
Posted
1 hour ago, Nilbobaggins said:

I assume that I will have to disassemble the regulator completely?

Whatever approach  you see as less risky and  feel more control over the task is right for you.

Regulator arm is made of spring and snapps back in place. 

Rgds

  • Like 1
Posted

Update: I fixed the regulator arm and the readings looks somewhat better.
However, I noticed something wierd: when I film the balance in slow motion the amplitude looks to be around 270° instead of the suggested ~140° from the timegrapher. I have only googled the lift angle (52°), but perhaps it could be a clue to what is wrong?

 

IMG_3006.jpeg

IMG_3012.jpeg

IMG_3011.jpeg

Posted
11 minutes ago, Nilbobaggins said:

when I film the balance in slow motion the amplitude looks to be around 270° instead of the suggested ~140° from the timegrapher

Just in case you have the same confusion I did for a long time: amplitude is measured as the rotation in each direction from the rest position of the balance.

So if you're seeing a total rotation of a running balance go 270 degrees in one direction then 270 degrees in the other, that's actually 135 degrees of amplitude, since it is measured from the center rest position of the balance.

Posted
11 minutes ago, Nucejoe said:

 You can  post a top view vid of the oscilator running on  youtube and gives  us its link here. Side view of the coil be nice too.

 

Sure:

Let me know if I should record anything else 🙂
 

Posted
4 hours ago, ManSkirtBrew said:

Just in case you have the same confusion I did for a long time: amplitude is measured as the rotation in each direction from the rest position of the balance.

So if you're seeing a total rotation of a running balance go 270 degrees in one direction then 270 degrees in the other, that's actually 135 degrees of amplitude, since it is measured from the center rest position of the balance.

No, 270° amplitude is 270° one way from the rest position, then 270° the other direction.

Posted (edited)
2 hours ago, mikepilk said:

No, 270° amplitude is 270° one way from the rest position, then 270° the other direction.

Yes, this is what I was saying, just differently.

If he's seeing 270 degrees of rotation from stop to stop, that's 135 degrees of amplitude, which matches what he's seeing on the timegrapher. The video he posted bears that out as well.

 

Edit for clarity: because the rest position should be halfway between the extremes.

Edited by ManSkirtBrew
  • Like 2

Join the conversation

You can post now and register later. If you have an account, sign in now to post with your account.
Note: Your post will require moderator approval before it will be visible.

Guest
Reply to this topic...

×   Pasted as rich text.   Restore formatting

  Only 75 emoji are allowed.

×   Your link has been automatically embedded.   Display as a link instead

×   Your previous content has been restored.   Clear editor

×   You cannot paste images directly. Upload or insert images from URL.


  • Recently Browsing

    • No registered users viewing this page.
  • Topics

  • Posts

    • I put the original broken glass (whats left of it) behind the acrylic. This is how it originally looked like. A cheap clock but I like it. Very noisy when the alarm comes wake up guaranteed! 🙂
    • If at all possible, find a service guide for the automatic movements your work on, because the lubrication procedures may have different requirements or rely on oils you would not use in a manual wind train (in addition to the braking grease you mentioned). Some autos like older Seikos do not have a manual wind option, so the procedure of letting down the mainspring without being able to use the crown may require a screwdriver in the ratchet wheel screw and great care. Do you have an auto movement you were planning to start with?
    • I am an amateur, so there's that. I do not get fixated on amplitude, lift angles, and beat error. However, 4.8ms would bug me if it were my watch. But you must judge your own skills to appreciate the possibility of going backward. I suggest, that you button it up let your friend enjoy the watch for now. As your skills progress, come back to it and correct it. I assume that this watch has a fixed hairspring pin. Some modern watches have an adjustable pin along with adjustable regulator. These are trivial to get in beat. I own a valjoux 726 my dad gave me on my 18th birthday (a looooong time ago). I broke the ratchet wheel with an aggressive wind 4 yrs ago. I have been waiting for my skills to progress before doing a service. I am close. Your advice is well placed and I will apply it.
    • I didn’t find any anomaly to the left of the red mark…reflection? this is the balance in its pivot in the inverted assembly. i can’t see any obvious kinks  and the spring is flat as far as I can see. Either the stud screw is missing, or it’s glued in… I don’t know. I’m loathe to fiddle with it. Any further insights? Thanks!
    • Update!  I've dismantled it, cleaned all the glue off, and rebuilt and lubricated the base movement. I'll leave the chrono part for another day. It's running well - great amplitude and keeping time, but it's got a beat error of 4.8ms.    How important is it to correct this? I'm worried that the potential for making things worse having to take the hairspring off and on repeatedly to adjust this. Would anyone here accept it at that?
×
×